View Full Version : Schuran Calcium Reactor


Kevin Douglas
11-07-04, 09:54
Anyone using the Schuran Jetstream 1 or 2 with a pH Probe and controller.

The probe fits in the tube just above the tap. I am getting some irregular readings and I am finding that the Co2 builds into a pocket in the tap area which gives false readings. Today it told me the reactor was running at 6.25 when it should never get that low becuase I have the Tunze Controller set at 6.5. When I release the probe to let out the gas the pH reading climbed to 6.9.
So this means the the reactor is running on a higher pH than I want it to ie:- 6.50
but the Controller thinks it is below 6.50 and keeps the Co2 switched off.

I think I have found a flaw in the design.
Anyone got any comments or found the same as me.

Atb

Tony B
04-10-07, 20:55
Yep I agree this is a design flaw; the reast of the reactor is superb but you need to mod it to correctly take a PH probe :) Sorry for the delay in responding :wave: :laugh:

Will
04-10-07, 22:04
you **** taker tony, taking his fix and then responding to this pmsl

Will
04-10-07, 22:11
just seen the other thread pmsl

Creepycoops
04-10-07, 22:30
Yep I agree this is a design flaw; the reast of the reactor is superb but you need to mod it to correctly take a PH probe :) Sorry for the delay in responding :wave: :laugh:

What is the mod as i am getting one very soon and intend to run a probe?

Cheers

Kev