There apperas to be an awfull lot of confusion about these terms and what they mean with relation to specific instruments, would be nice if we could get a collective understanding and put something usefull in the FAQ section.

My fave related quote "kim "

If people don't know to which index of SG their instrument (hydrometer or refractometer) is calibrated, they can't possibly know their density to an accuracy any better than +/- 0.004 g/cc approx. So measuring to 0.001 is really silly.

That uncertainty is +/- 5 % on the salinity scale (we are targetting about 35 %), so it is quite considerable. Could easily lead to low alkalinity, for example.

Haven't heard of many tanks crashing for this reason, but fusspots like me would sleep a little less well. My moral....know your SG first ! Targeting an SG without knowing which one the instrument reads is careless."

In general, if you don't know your salinity then it's pretty much pointless trying to get other paramaters into a particular range.

A snippet to bring the point home from AndyH

The really bad thing about hypo-salinic tanks (ie. less that 1.0264 @ 25C) is that it screws up all the other measurements for example Ca, alk, Mg, iodine (in its various forms) etc etc etc and people are seemingly obsessed with having 400 ppm Ca in their tanks whereas at 31 ppt (S=31) one would only expect the Ca level to about 350ppm.If you want to shoot for 31ppt then at least expect all the ions to be 12% under normal values.



Firstly, typicall instruments

Hydrometers
-----------

Sea Test - Swing arm
deep six - Swing arm
TMC Glass Jobbie
e.t.c

Kim on Hydrometers


It is to be used at 25 C (ie that is the temperature at which it is calibrated). The reading is a specific gravity, and in this case the SG is computed as the density of the sample divided by the density of water at 4 C. This is the most common definition of SG quoted in relation to oceanographical work, apparently.

Since the density of water at 4 C is pretty much 1 g/cm, the SG measured by the hydrometer is, in fact, close to being numerically equal to the density of sample.

If you want to be fussy, then because the density of water at 1 atm and 4 C is in fact 0.99997 g/cm, to get the true density just subtract 0.00003 g/cm from the hydrometer reading. But I would guess that this is finer than the instrument is capable of measuring so it's easy to see why this was described as a density hydrometer.

simon on Hydrometers

"Density is as its says density of water.

SG is an expression of density in relative to pure water at a reference temperature. Natural seawater has a SG of 1.024-1.026 and is equivalent to 35ppt salinity

If you are saying your reading by that meter is 1.024 I would say your sg is approx =1.028

If you had said it was 1.026 then would be sg 1.030 aprrox

My meter reads same as my iks probe set to density when both say 1.022 my salinity is 35ppm again by probe."



simonh "Using the TMC hydrometer at 25C a specific gravity of 1.0264 is equivalent to a natural seawater salinity of 35 PSU."



Can we come up with a simple instruction set and a set of reference tables that will allow users to use the above instruments within reason?


Refractometers?

It is an ATACO seawater model (not saltwater=brine) model? Salt water refractometers read about 0.15 "units" out vs seawater. Atago S/Mill-E and the ATC S/Mill-E (ATC = temperature compensated) at £200 is a tad expensive though.

the general blurb "RHS-10ATC dual scale portable Salinity Refractometer in a blue, foam-lined hard case. The RHS series of portable refractometers is designed to measure the salinity of sea water and saline solutions of similar concentration. One scale checks the NaCl levels with the range of 0-100 ppt (with 1 ppt scale divisions) and the other scale gauges Specific Gravity with a range of 1.000 to 1.070 (+/- 0.001 accuracy). Both enable the direct determination of salinity in water that contains dissolved salt and little or no other dissolved solids.

Hydrometers are nice, but they can be inaccurate. The RHS-10ATC is designed to be very accurate to protect your investments!

All RHS models use ambient light, no battery or power source is required, making them truly portable. Models with the “ATC” suffix are equipped with “Automatic Temperature Compensation” for accurate measurements without recalibration after shifts in ambient working temperature (field use).

Included are: a pipette for dropping test fluids onto the prism (never dip the refractometer into the test fluid!), a mini screwdriver for scale calibration, and a Users Guide"


As a general rule, can we agree on the values that will represent s=35 (35ppt) on a brine refractometer (as it's what most have).





Electronic Gizmos

IKS

AndyH "I set the reading to salinity rather than density so have no idea if that causes problems. Since both density and SG are temperature dependent I don't see the point in either of those units of measure. "
andyh "Similarly with conductivity. NSW should be 53mS/cm. "

So, if using an IKS, stick with reading your measurements in Salinity!


Tunze, the reference tables supplied by Tunze are inaccurate ?


Kim provides a more accurate table

If you convert your SG to density, then for "standard" seawater at 25 C:-

Density...........Conductivity.........Salinity

1.01207..............32.17.................20
1.01582..............39.26.................25
1.01957..............46.21.................30
1.02334..............53.02.................35
1.02712..............59.74.................40

Density (g/cm3), conductivity (mS/cm), salinity (g/kg).

(taken from Kaye and Laby, Physical and Chemical Constants, 14th Edition).

this is close to the IKS supplied table.

If using a Tunze, use the table above.


No real issues in this area. i.e 53ms/cm = s=35.




What is SG

by Kim ?

My point about SG is that there are many definitions of it. I can't remember them all, but let me illustrate with an example, rather than waving arms and confusing myself too..

Definition 1

SG is the density of the sample divided by the density of pure water at 4 C. (Density of pure water at 4 C is 1.0000 g/cc (almost).)

Definition 2

SG is the density of the sample divided by the density of pure water at 25 C.
Density of pure water at 25 C is 0.9970 g/cc)

Assume that you have a hydrometer (or refractometer) which measures the SG of water at 25 C, and that your water has a salinity of 35 %. That is a density of 1.0233 g/cc at 25 C. Note that the temperature at which a hydrometer works correctly is usually not the "Definition Temperature" - it's just a result of the design.

Hydrometer 1 will read 1.0233, hydrometer 2 will read 1.0264. They are both correct - they are measuring different things. So it's imporatnt to know what they measure ! If you don't know this, arguing about 0.001 is just not useful.

Hope that helps. If nothing else, hope it suggests that SG is just a ridiculous concept anyway. Invented by brewers on a Friday afternoon....


what is Salinity?

kim

ppt is weight per weight, but if you want to be fancy, seawater salinity is quoted as (eg) 35, not 35 ppt, and it is no longer defined in terms of weight per weight but electrical conductivity relative to a chemical standard. Hence no units....it is like a percentage. Not a lot of people know that......Even fewer care.)






what's the point by Glenn.

I think the problem i have with the debate over the benifits of a refractometer are the principles of the equipment.

Basically any measuring instrument is made to a degree of accuracy and precision, typically related to the cost of the kit.

Accuracy is how close to a measured value the instrument can get and precisiosn being the ability to reproduce the results.

Simply being a refractometer does not mean an insturment is good at both these concepts.

It may be that the refractomer suffers from fewer inherent problems than the swinarg thing but it does not necessarily mean its either more accurate or precise because its a refractometer.

For the record i dont think that swing arm devices are particulalry well made, accurate or precise.

I do think that a hydrometer has every chance of being both accurate and precise and there is no reason why they should be discounted as a reliable means for measuring the specific gravity of out tanks.

Finally i do wonder what we are going to do with the degreee of accuracy that we often see spoken of.

IF you have a 150 gallon system like mine, the top off adds roughly 500ml per addition, so i dont suppose the sg varies much. If i didnt have an auto top off i would add around 2 gallon per day and the percentage change would be larger.

If you translate these kind of issues to smaller tanks the percentage change over a day can be quite dramatic in the.

So why do we need to worry if we measure the sg to within say 0.001 ?


i.e. even if we have an accurate measure, what can we do with it, if it can't be used reliably to maintain s=35 then what's the point?

and finally

andyh "About the only thing we can rely on is a calibrated (with certificate) hydrometer and thermometer."



So once you've all got Kim and Gleen to translate the above into something usefull we might finally have a single reference point for the future :whistling: :D

over to you

Kev s
29-09-05, 16:44
When is series of pocket guides coming out??

Kev

~Tony~
29-09-05, 17:29
About the only thing we can rely on is a calibrated (with certificate) hydrometer and thermometer
I agree with this. I obtain my Salinity reading using a Lab grade glass density hydrometer and Lab grade thermometer from Brannan.

I then use this table to lookup my Salinity from these 2 readings.

Density of sea water . . . . . . . . . . . . . . . . . . . Salinity
temp .. 34.0 ... 34.1 ... 34.2 ... 34.3 ... 34.4 ... 34.5 ... 34.6 ... 34.7 ... 34.8 ... 34.9 ... 35.0 ... 35.1 ... 35.2 ... 35.3 ... 35.4 ... 35.5
25.0 .. 22.588 22.263 22.739 22.814 22.890 22.965 23.041 23.116 23.192 23.267 23.343 23.419 23.494 23.570 23.645 23.721
25.1 .. 22.557 22.633 22.708 22.784 22.859 22.935 23.010 23.086 23.162 23.237 23.313 23.388 23.464 23.539 23.615 23.691
25.2 .. 22.527 22.602 22.678 22.753 22.829 22.904 22.980 23.055 23.131 23.207 23.282 23.358 23.433 23.509 23.584 23.660
25.3 .. 22.496 22.572 22.647 22.723 22.798 22.874 22.949 23.025 23.100 23.176 23.252 23.327 23.403 23.478 23.554 23.629
25.4 .. 22.466 22.541 22.617 22.692 22.768 22.843 22.919 22.994 23.070 23.145 23.221 23.296 23.320 23.447 23.523 23.599
25.5 .. 22.435 22.511 22.586 22.662 22.737 22.813 22.888 22.964 23.039 23.115 23.190 23.266 23.341 23.417 23.492 23.568
25.6 .. 22.405 22.480 22.555 22.631 22.706 22.782 22.857 22.933 23.008 23.084 23.159 23.235 23.310 23.386 23.461 23.537
25.7 .. 22.374 22.449 22.525 22.600 22.676 22.751 22.827 22.902 22.977 23.053 23.128 23.204 23.279 23.355 23.430 23.506
25.8 .. 22.343 22.418 22.494 22.569 22.645 22.720 22.796 22.871 22.947 23.022 23.097 23.173 23.248 23.324 23.399 23.475
25.9 .. 22.312 22.387 22.463 22.538 22.614 22.689 22.765 22.840 22.915 22.991 23.066 23.142 23.217 23.293 23.368 23.444
26.0 .. 22.281 22.356 22.432 22.507 22.583 22.658 22.734 22.809 22.884 22.960 23.035 23.111 23.186 23.262 23.337 23.413
26.1 .. 22.250 22.325 22.401 22.476 22.552 22.627 22.702 22.778 22.853 22.929 23.004 23.080 23.155 23.230 23.306 23.381
26.2 .. 22.219 22.294 22.370 22.445 22.520 22.596 22.671 22.747 22.822 22.897 22.973 23.048 23.124 23.199 23.275 23.350
26.3 .. 22.188 22.263 22.338 22.414 22.489 22.564 22.640 22.715 22.791 22.866 22.941 23.017 23.092 23.168 23.243 23.319
26.4 .. 22.156 22.232 22.307 22.382 22.458 22.533 22.608 22.684 22.759 22.835 22.910 22.985 23.061 23.136 23.212 23.287
26.5 .. 22.125 22.200 22.276 22.351 22.426 22.502 22.577 22.652 22.728 22.803 22.879 22.954 23.029 23.105 23.180 23.255
26.6 .. 22.093 22.169 22.244 22.319 22.395 22.470 22.545 22.621 22.696 22.772 22.847 22.922 22.998 23.073 23.148 23.224
26.7 .. 22.062 22.137 22.213 22.288 22.363 22.439 22.514 22.589 22.665 22.740 22.815 22.891 22.966 23.041 23.117 23.192
26.8 .. 22.030 22.106 22.181 22.256 22.332 22.407 22.482 22.558 22.633 22.708 22.747 22.859 22.934 23.010 23.085 23.160
26.9 .. 21.999 22.074 22.149 22.224 22.300 22.375 22.450 22.526 22.601 22.676 22.752 22.827 22.902 22.978 23.053 23.128
27.0 .. 21.967 22.042 22.117 22.193 22.268 22.343 22.419 22.494 22.569 22.645 22.720 22.795 22.870 22.946 23.021 23.097

Where 22.588 means 1.022588 g/ml etc
Obtained from Seawater Density Calculator (http://www.es.flinders.edu.au/~mattom/Utilities/density.html). The figures match the table from Kim's book accurately enough for me to trust the calculator.

Ignore SG. Use calibrated instruments. Obtain salinity value. Only way to go for me :)

Tony